Association closed immediately by Konica ultrasound #832
Replies: 6 comments 2 replies
-
Just to follow up on this issue, as far as I know, Konica has accepted that there could be a bug on their end. |
Beta Was this translation helpful? Give feedback.
-
Moved to discussions to keep this open as information. |
Beta Was this translation helpful? Give feedback.
-
For the record, in case future parties stumble across this discussion, we heard back from Konica
After field testing, it was found that allowing the PDU to be configurable and setting a 16k PDU size worked correctly with the HS2 unit. |
Beta Was this translation helpful? Give feedback.
-
Describe the bug
An Association cannot be established with a Konica ultrasound HS2 unit.
Expected behavior
The Association should not immediately close after being accepted.
Steps To Reproduce
Actual code is close to this but has more services (DWML service, async DICOM Push event handling, async Query/Retrieve handling) that were elided, but I tried to keep the honest look of the code as best I could:
Your environment
We do not have direct access to the user's workstation, but the following are the case:
Windows 2010
Python 3.7
pydicom 2.2.0
pynetdicom 2.0.0
NOTE: this is compiled by pyinstaller.exe and installed on Windows via NSIS
Beta Was this translation helpful? Give feedback.